What is the torque guideline for a Astra 3.0? Is it 15 newton centimeters?
 
rkm rdt

rkm rdt

Well-Known Member
Full Member
Messages
21,447
Reaction score
3,288
Did it break?

Whatever you do,don't use the stock 15 degree tapered abutment.It is 1/10 micron thick at the base of the wall and will fracture!
 
ParkwayDental

ParkwayDental

Well-Known Member
Full Member
Messages
1,774
Reaction score
251
Well the abutment is fine. The Doc called me today and he said the screw broke. I asked him what he torqued it to and he said 30. I want to say though that it is suppose to be torqued to 15 on the the 3.0?
 
k2 Ceramic Studio

k2 Ceramic Studio

Well-Known Member
Full Member
Messages
1,084
Reaction score
39
What is the torque guideline for a Astra 3.0? Is it 15 newton centimeters?
I am taking a guess but that sounds way too low, hand tightened is 10 so I think your 15 is probably the setting for temp abutments or healing caps. Will have a look in the old Astra book now and see what it says.
 
k2 Ceramic Studio

k2 Ceramic Studio

Well-Known Member
Full Member
Messages
1,084
Reaction score
39
OsseoSpeed™

TiDesign™ abutments are manufactured with a hex for indexing and are packaged as a kit with a final screw. TiDesign™ stock titanium abutments are available for all three (3.0 yellow, 3.5/4.0 aqua, 4.5/5.0 lilac) implant/abutment interface connection groups.

3.0 TiDesign™ abutment
3.0 TiDesign™ abutments are machined with a emergence profile of 4.0mm and a choice of collar heights of 1mm, 2mm and 2.4mm. The 1mm and 2mm collar height abutments are straight and the 2.4mm collar height abutment has a 15 degree angle machined in relation to the flat side of the hex. Number markings indicate the emergence profile and a single black line indicates the 3.0 implant/abutment connection group.

Emergence profiles
Straight 3.5/4.0 TiDesign™ abutments are available with a choice of three emergence profiles (4.5, 5.5, 6.5mm) and a collar height of 1.5mm or 3mm for each emergence profile. Two angled abutments with a angulation of 20 degrees and a collar height of 3mm with a choice of two (4.5, 5.5mm) emergence profiles are also offered. Number markings indicate the emergence profile and a double black line indicates the 3.5/4.0 implant/abutment connection group.

Collar height
4.5/5.0 connection diameter TiDesign™ abutments are machined with two (5.5, 6.5mm) emergence profiles and either a 1.5mm or 3mm collar. One 20 degree angled abutment with a emergence profile of 5.5mm and a collar height of 3.5mm is also available. Number markings indicate the emergence profile and the absence of a black line indicates the 4.5/5.0 implant/abutment connection group.

Collar height
Direct Abutment API™ titanium abutments are conceptually designed for posterior restorations and intended to be installed by the surgeon. These abutments are machined as a one piece unit (abutment body and screw) without a index (hex) at the implant/abutment connection interface. Intended for abutment level impressions with appropriate plastic impression copings.
3.5/4.0 platform group direct abutments are available with a 4mm emergence profile and a choice of four (0.5, 1, 2.5, 4mm) collar heights or a emergence profile of 5mm and three (1, 2.5, 4mm) possible collar heights. 4.5/5.0 platform group direct abutments are available with a 5mm emergence profile and a choice of four (0.5, 1, 2.5, 4mm) collar heights or a emergence profile of 6mm and three (1, 2.5, 4mm) possible collar heights.

Emergence profile
All abutments are marked with a black line 1mm from the top of the abutment in case the abutment needs intraoral height adjustment. Analogs are marked in the same manner and can then be reduced to the same height, using the marker as a reference point. A number printed on the abutment indicates the emergence profile (4, 5 or 6mm). These abutments are usually sold as a kit containing the abutment, abutment carrier, impression coping, analog and burnout copings but also available a la carte. Abutments are torqued to 25Ncm regardless of platform connection diameter group.

Collar height
ZirDesign™ abutments are made through and through in zirconia, machined in a similar geometry as TiDesign™ abutments. The 3.5/4.0 platform group is offered with a choice of two (1.5, 3mm) collar heights and two (4.5, 5.5mm) emergence profiles for either collar height. A pre angled abutment with a 20 degree angle, 5.5 emergence profile and

Emergence profile
collar height of 3.5mm is part of the zirconia abutment product line. The 4.5/5.0 platform group is also offered with a choice of two (1.5, 3mm) collar heights and two (5.5, 6.5mm) emergence profiles for either collar height. A pre angled abutment with a 20 degree angle, 5.5 emergence profile and collar height of 3.5mm is part of the zirconia abutment product offering. A final screw is included with each abutment as a kit.

UCLA
CastDesign™ abutments are UCLA's made with a hexed, machined gold implant/abutment connection interface and a plastic cylinder pressed on to the gold portion of the abutment. Each platform group (3.5/4.0, 4.5/5.0) has its own geometry and made to fit only their respective platform group. A final screw is included with each abutment as a kit. Stock, non indexed UCLA's are not available for Astra Tech implants.

TempDesign™
TempDesign™ abutments for each implant/abutment connection interface group (3.5/4.0, 4.5/5.0) are manufactured with a titanium base and a plastic cylinder pressed on to the titanium base. Abutments are available with and without indexing feature (hex). Through and through temporary titanium with and without indexing feature (hex) are also offered. A final screw is included with each abutment as a kit. The manufacturers recommended torque setting is 15 Ncm regardless of platform connection diameter group.

UniAbutments serve as intermediary abutments for splinted, screw reatained restorations and substructures. Theses abutments may be utilized in situations with exessive tissue height (≥2mm) above implants or to correct a unfavorable final screw access path. Special UCLA abutments (discussed below) or CAD CAM substructures are required for UniAbutments.

UniAbutment straight
Straight UniAbutments are non indexed and machined as a mono block, integrating the screw and abutment in to a single entity, either with a 20 degree or 45 degree cone to achive a common path of draw for the final restoration. The correction of draw may be up to 40 and 90 degrees respectively. Both platform diameter (3.5/4.0 aqua, 4.5/5.0 lilac) UniAbutments are not color coded for platform identification and available in six collar heights of 0.5, 1, 2, 4, 6 and 8mm. A special UniAbutment driver is required to torque (20Ncm) the abutments in place.

UniAbutment angled
Angled UniAbutments for 3.5/4.0 and 4.5/5.0 platform implants are machined with a 20 degree angle as a two component set (screw and abutment). Offered as indexed and non indexed versions with a collar height of 0.8mm and 2mm. The abutments have a height of 5.1mm plus the collar height, making the stack height of the components (UniAbutment plus UCLA or CAD substructure) considerable. No color coding or other markings are present on the abutments for platform diameter identification.

UniAbutment UCLA's
UniAbutment UCLA components are machined to fit the type of UniAbutment (straight 20°, straight 45°, angled) regarless of platform diameter (3.5/4.0, 4.5/5.0). UCLA's are made with a non hexed, machined gold, plastic or titanium implant/abutment connection interface. Gold and titanium interfaces are press fit with a plastic cylinder on to the metal portion of the abutment. A final screw is included with each abutment as a kit and torqued to 15Ncm.

OsseoSpeed™ Profile

Profile TiDesign™
Profile TiDesign™ abutments are manufactured with a hex for indexing and are packaged as a kit with a final screw. Straight 4.5/5.0 TiDesign™ abutments are available with a choice of two emergence profiles (5.5, 6.5mm) and a collar height of 1.8mm or 3mm for each emergence profile. One 15 degree angled abutment with a emergence profile of 6.2mm and a collar height of 2.7mm is also available. Number markings indicate the emergence profile and a "P" indicates the Profile 4.5/5.0 implant/abutment connection group.

Profile Temporary
abutment
One straight engaging temporary abutment is currently available for the Profile implant product line. The abutment has a 1.5mm sloping collar and the lowest point of the slope synchronized with the flat side of the indexing feature (hex). A total height of 9mm and a emergence profile of 5.3mm. The final screw is included with each abutment as a kit.
 
Last edited:
k2 Ceramic Studio

k2 Ceramic Studio

Well-Known Member
Full Member
Messages
1,084
Reaction score
39
Just looked in the Astra product catalog for 2012 and it says TiDesign 3.0 Titanium, Recommended Torque:Yellow=15Ncm (Aqua=20Ncm) (Lilac=25Ncm) Looks like your guy over torqued it.
 
Last edited:
GoldRunner

GoldRunner

Active Member
Full Member
Messages
136
Reaction score
26
I have had doctors tell me that the screw also "broke" and that it was something that we did or the manufacture. Upon looking at the seating verification x-ray is was evident the the abutment was never properly seated. It remains tight for a while and then loosens as the abutment drops into place. This usually occurs because they use the screw to seat the abutment and it jams itself into a slightly incorrect place (there are too many indexing positions). The abutment must be seated by hand, and then the screw is introduced.

They might notice a slight rotation to the restoration, but it is usually very minimal. Now that the screw is loose, the abutment drops down slightly and easily breaks the screw when loaded. While this might not be what happened in your case (Dr. over torqued) this is a very real problem. It is further exacerbated by the problem that the same thing occurs during placement of the impression coping. Your crown will look fine on the model and then be slightly rotated in the mouth. Another remake!

This is really a major design flaw and is amplified by the lack of clinical support provided by Dentsply. If the Drs are informed about the problem they will do it "their way" anyhow.
 
rkm rdt

rkm rdt

Well-Known Member
Full Member
Messages
21,447
Reaction score
3,288
After 3 fractures,the rep disappeared and will not honour the warranty.

They have lost me as a customer!!!!!!!!!
 
actittle

actittle

Señor Member
Full Member
Messages
362
Reaction score
74
I have had doctors tell me that the screw also "broke" and that it was something that we did or the manufacture. Upon looking at the seating verification x-ray is was evident the the abutment was never properly seated. It remains tight for a while and then loosens as the abutment drops into place. This usually occurs because they use the screw to seat the abutment and it jams itself into a slightly incorrect place (there are too many indexing positions). The abutment must be seated by hand, and then the screw is introduced.

They might notice a slight rotation to the restoration, but it is usually very minimal. Now that the screw is loose, the abutment drops down slightly and easily breaks the screw when loaded. While this might not be what happened in your case (Dr. over torqued) this is a very real problem. It is further exacerbated by the problem that the same thing occurs during placement of the impression coping. Your crown will look fine on the model and then be slightly rotated in the mouth. Another remake!

This is really a major design flaw and is amplified by the lack of clinical support provided by Dentsply. If the Drs are informed about the problem they will do it "their way" anyhow.

This is one of the many reasons we always fabricate an abutment placement jig for the docs to use for seating the abutments on our cases.
 
lcmlabforum

lcmlabforum

Well-Known Member
Full Member
Messages
1,476
Reaction score
160
I have had doctors tell me that the screw also "broke" and that it was something that we did or the manufacture. Upon looking at the seating verification x-ray is was evident the the abutment was never properly seated. It remains tight for a while and then loosens as the abutment drops into place. This usually occurs because they use the screw to seat the abutment and it jams itself into a slightly incorrect place (there are too many indexing positions). The abutment must be seated by hand, and then the screw is introduced.

They might notice a slight rotation to the restoration, but it is usually very minimal. Now that the screw is loose, the abutment drops down slightly and easily breaks the screw when loaded. While this might not be what happened in your case (Dr. over torqued) this is a very real problem. It is further exacerbated by the problem that the same thing occurs during placement of the impression coping. Your crown will look fine on the model and then be slightly rotated in the mouth. Another remake!

This is really a major design flaw and is amplified by the lack of clinical support provided by Dentsply. If the Drs are informed about the problem they will do it "their way" anyhow.
I don't think they adjusted the interprox contacts fully before trying to seat the crown.
Orientation jig helps, but if prox contacts are preventing it from fully seated, it does not
matter what torque was applied, something will give - either loosening and fracture or
fracture before it loosens.
If you don't take the time to do it right the first time, when will you find the time to
redo it? - was something my colleague in Kansas City always tell the students.
Not sure you want a client like this anyway . . .