They're running on 220. Hold times are 2 hours.
This setup was working fine for years with no issues, which makes it puzzling. I can't pinpoint to anything substantial that would have changed.
As for temps, I re-checked on more runs (I must have made measuring mistakes before or used faulty rings) - all started from close to room temperature - and I now have:
This setup was working fine for years with no issues, which makes it puzzling. I can't pinpoint to anything substantial that would have changed.
As for temps, I re-checked on more runs (I must have made measuring mistakes before or used faulty rings) - all started from close to room temperature - and I now have:
- Bottom layer 15-20°C lower than the middle layer
- Middle layer on target
- Top layer < 5°C above the middle layer